I recently ordered the SCT-013-000 but received SCT-013-000V by mistake. From the product range I see this is the difference:
SCT-013-000 Primary AC current: 100A Secondary current: 50mA
SCT-013-000V Primary AC current: 100A Secondary voltage: 1V
Can I remove the two burden resistors within the SCT-013-000V to make the sensor into an SCT-013-000? I.e. from VT to CT; what about the TVS, is the same TVS inside the CT and VT models?
You can remove the burden within the -100V, but you really need to have the TVS - because there wonât be one in the -100V while it has the internal burden (or if there is, there is no need for one â meaning I donât wholly believe the YHDC data sheet).
If there is a TVS present, leave it there. If there isnât, you need a bipolar TVS, or a pair of zener diodes, in series but facing opposite directions (i.e. A-A or K-K). Almost any voltage will do, 5.6 V upwards to say 12 V - the sole purpose is to give the c.t. a load to hold the voltage down if itâs disconnected from the external burden, the one inside the emonTx, emonPi or on your breadboard.
It wonât exactly be a SCT-013-000, because I think the âvoltageâ versions have a slightly different number of secondary turns (1860 vs 2000 nominally), youâll need to adjust the calibration to take account of that. I think the ratio will be 100 A : 53.76 mA
if you can measure the internal burden after youâve got it out, Iâd be interested to know what the value is.
I presume youâve figured out how to take the c.t. apart â there are two plastic âspring clipsâ that hang over the edge of the bobbin, push those aside (this needs 4 hands) and the core and bobbin will slide out.