CRC Error - Sontex Supercal 531 Integrator

After a power cut in today’s storm my Sontex Supercal 531 heat meter integrator has decided to no longer work!

Basically the heat meter is now displaying a “CRC error” message in the display and I cannot bring it back to life. It’s the mains powered unit with battery back up. I’ve tried to reset it multiple times by powercycling the display module on the front, killing the mains power and removing the battery jumper on the main board, but all to no avail. Presumably there’s some kind of corruption in the EEPROM register that’s preventing it from coming back to life.

Does anyone have experience with this problem and a possible solution? I think these units have software to reprogram them (Prog531 and an optical head module), but I don’t have access to this and I’m not sure if this is a recoverable situation anyway.

Any advice would be most appreciated.

I’ve heard of a few people having this issue. Apparently, it’s a memory corruption issue which is unrepairable. Replacement will probably be required, I’ve not heard of any meters being recovered from this error. Please let us know if you have any luck :crossed_fingers:

I’ll reach out to the supplier to see if there’s any “service back-end” recovery - as you say probably unlikely. I understand the lockdowns in the meters, if used in a billing environment, but for the rest of us just using them to feed into monitoring it’s annoying that there’s no way to easily reload firmware etc. Will let you know what happens. Probably end up buying a secondhand replacement.

Be careful with 2nd hand sontex meters, the flow measument method that Sontex use is quite susceptible to dirt, if the meter is dirty inside it will quietly under report flow rate and therfore heat without any notification.

Even though ultrasonic meters are more susceptible to air than sontex at least they will display an error if they can’t measure flow rate. Ultrasonic meters are less susceptible to dirt.

If you require glycol calibration then Sontex is your only choice.

My thought is just to change the actual integrator unit. The installed flow meter is the Sontex Supermatic 440 and that would just stay in place and the integrator unit would be swapped out.

I do run with glycol and the original 531 was configured accordingly. I did notice a while back, that the 531 on its MBus feed gave both the non-corrected and corrected glycol heat power outputs simulataneously. If I got a used 531 integrator that hadn’t been configured for glycol I could just create a feed into emoncms that has the correction factor applied instead. At least that’s my thinking - hope that makes sense?

I’m not sure about the 531, but on other Sontex meters the flow meter is paired to the integrator, so swapping it may not be possible. We have the Sontex programming software to be able to apply glycol calibrations to Sontex 789 and Supercal5. I’m not sure about the 531, we’ve never been asked to program one since they have been discontinued a few years ago.

If you do decide to replace it we have stock of Sontex 789:

In case it’s useful to know, I recently sent my Superstatic 449 (also discontinued) to DMS Ltd for recalibration from glycol to water. They may be able to advise on how to repair or swap a 531.

Thanks Glyn and Tim - I will follow up and advise on this when I’ve spoken to DMS and figured out the best option.

I still think that a £500+ metering set-up which packs up after 4 years isn’t great…

FWIW sometime you can clip the calculator face of a failed 531 onto a good baseplate and have it recover the eeprom error.

@glyn.hudson check email on 531

Thanks Marko.

The update is as follows…

I sent the meter back to Chris at DMS and he managed to bring it all back to life with retained data. Already returned to me and wired back in and I can confirm all is well.

It turned out that during commissioning I inadverently forgot to remove a plastic tab with an orange arrow from the reverse of the front calculator module. A common mistake I’ve discovered! Removal activates the back up battery (CR2032) on the front calculator and provides power for EEPROM writes during power failure. Living out in the sticks I’ve encountered many power cuts over the last four years, but I’ve been lucky as a write wasn’t taking place at the moment power died, until last Friday!

I’m not sure if the back up battery tab is the same on a different meter like a Supercal 5I, but worth checking to be sure to avoid this kind of problem. Chris explained that the odds of recovery are 50/50 and I got away with it. He provided excellent service and immediate turnaround which I’m grateful for and I have to highly recommend DMS. Anyone else with a similar problem should just reach out to him.

Back now to getting my fix of heat pump monitoring!!

That’s great to hear! Yes, DMS are excellent. We use them as a suppler for Sontex meters.

For a £500 piece of hardware, that’s disappointingly fragile firmware :frowning: Either an A/B partition or recovery to default state would salvage it. I doubt many people would find someone to repair it.