Hi,

I have two emonTHs running, one with an optical sensor for reading electricity led pulses and the other which reads a reed sensor for gas.

I’ve left both running for an extended period of time and I have had to periodically adjust them both however given that I have had to do this I am concerned about the historic data and wanting to fix this so that this doesn’t happen again.

For both monitors I have created virtual feeds which should represent the meter’s current display, it is the offset that I have had to to keep adjusting to bring the live figure back into line.

How best can I go about identifying and rectifying the cause? I believe the gas meter reading with the reed sensor is due to switch bounce but I’m not sure what would be causing the electric meter reading.

The meters I have are as follows:

- Llandis+Gyr E110
- Estler BK-G4M

Below I have attached how I calculate the inputs/outputs. Any advice would be greatly appreciated as I’m currently being over charged by £60 by my current electricity supplier due to a mistake made by the meter reading person so would like to not have any future concerns about the readings reported by Emoncms.

**Electric**  
Pulse input calculation:

 ![image](https://community.openenergymonitor.org/uploads/default/original/3X/f/d/fd41ece9b937fde3460cad3c7f80a482f0a24b9e.png)

Virtual reading:

 ![image](https://community.openenergymonitor.org/uploads/default/original/3X/f/5/f578b8e1e92d1d5f66cfabd9b2e1b381c94c91c0.png)

**Gas**  
Pulse input calculation::

 ![image](https://community.openenergymonitor.org/uploads/default/original/3X/f/a/fa917ad11f232292537ae83dc7047ca931040331.png)

Virtual reading:

 ![image](https://community.openenergymonitor.org/uploads/default/original/3X/1/6/162f63ea8e06aead786ac6c92129c2a40e38987d.png)

Thanks in advance.

---

<div class="post-metadata">

**Author:** ![Robert.Wall](https://community.openenergymonitor.org/user_avatar/community.openenergymonitor.org/robert.wall/32/38246_2.png) [@Robert.Wall](https://community.openenergymonitor.org/u/Robert.Wall)\
**Post date:** [5 April 2022 22:54 UTC](https://community.openenergymonitor.org/t/readings-off/20176/2 "2022-04-05T22:54:37Z")

</div>

I had a magnetic reed switch connected to an emonTH for my gas meter, that counted approximately 2.5 per pulse. The addition of a 0.1 µF capacitor across the reed switch completely eliminated the problem. I suggested adding a note about this to the Wiki in January - it’s not happened.

I also have an optical pulse sensor, but that’s connected to an emonPi; so not directly relevant. That has on occasions picked up a nearby fluorescent lamp (with a thermal starter) being switched on or off. Others have found light getting to the sensor has been counted (perhaps obvious in hindsight).

---

<div class="post-metadata">

**Author:** ![Gwil](https://community.openenergymonitor.org/user_avatar/community.openenergymonitor.org/gwil/32/29028_2.png) [@Gwil](https://community.openenergymonitor.org/u/Gwil)\
**Post date:** [7 April 2022 22:17 UTC](https://community.openenergymonitor.org/t/readings-off/20176/3 "2022-04-07T22:17:18Z")

</div>

> [@Robert.Wall](#):
>
> I suggested adding a note about this to the Wiki in January - it’s not happened.

Sorry about that, Robert. I can add that note for you.

> [@Robert.Wall](#):
>
> The addition of a 0.1 µF capacitor across the reed switch completely eliminated the problem [of getting ~2.5 counts per each individual pulse]

---

<div class="post-metadata">

**Author:** ![Robert.Wall](https://community.openenergymonitor.org/user_avatar/community.openenergymonitor.org/robert.wall/32/38246_2.png) [@Robert.Wall](https://community.openenergymonitor.org/u/Robert.Wall)\
**Post date:** [7 April 2022 22:47 UTC](https://community.openenergymonitor.org/t/readings-off/20176/4 "2022-04-07T22:47:10Z")

</div>

Yes please, I think it would be helpful.

---

<div class="post-metadata">

**Author:** ![NotVeryOriginal](https://community.openenergymonitor.org/letter_avatar_proxy/v4/letter/n/c4cdca/32.png) [@NotVeryOriginal](https://community.openenergymonitor.org/u/NotVeryOriginal)\
**Post date:** [10 April 2022 20:46 UTC](https://community.openenergymonitor.org/t/readings-off/20176/5 "2022-04-10T20:46:28Z")

</div>

Robert, when you say across you mean like this?

![image](https://community.openenergymonitor.org/uploads/default/original/3X/8/9/893ea739d1ffe7d30357589262d21ffd02ba1425.png)

---

<div class="post-metadata">

**Author:** ![Robert.Wall](https://community.openenergymonitor.org/user_avatar/community.openenergymonitor.org/robert.wall/32/38246_2.png) [@Robert.Wall](https://community.openenergymonitor.org/u/Robert.Wall)\
**Post date:** [10 April 2022 21:02 UTC](https://community.openenergymonitor.org/t/readings-off/20176/6 "2022-04-10T21:02:31Z")

</div>

Something like that - you haven’t drawn a capacitor. 😆 It should look like this:

![bitmap](https://community.openenergymonitor.org/uploads/default/original/3X/4/0/40b4e964e529eec8cc204fe345c602b69115f5d0.png)

---

<div class="post-metadata">

**Author:** ![NotVeryOriginal](https://community.openenergymonitor.org/letter_avatar_proxy/v4/letter/n/c4cdca/32.png) [@NotVeryOriginal](https://community.openenergymonitor.org/u/NotVeryOriginal)\
**Post date:** [5 September 2022 20:32 UTC](https://community.openenergymonitor.org/t/readings-off/20176/7 "2022-09-05T20:32:20Z")

</div>

Robert, thanks for the suggestion of using a capacitor, it worked!