Good news for a change

I thought it would be good to share the solution I adopted for calibrating my recently purchased EmpnPi. After a fair bit of faffing about I managed to get the system working but the live power figures were reading about 10% high. I looked at the AC mains figure and that was also 10% high compared to my fluke DMM so I inserted a resistive divider to allow the measured AC to be adjusted to equal the DMM and now everything is spot on.

Happy bunny now.

1 Like

You could have altered the “scales = …” values in emonhub.conf to achieve the same result.

Excellent! thank you Robert, I shall give that a go.

Changed from 0.01 to 0.0095. Now much more accurate, thank you.

Bear in mind that the maths to calculate power is done a long time before the data reaches emonHub, so you must apply the same scaling to the powers as well as the voltage.

1 Like